If this value is higer then roughly 130 or so, and the air (28/69) ration is good and all other tune parameters look OK, it means there is a high chemical background in the analyzer. Best thing to do is to bake out the source at 300C and quad at 200C overnight. I like to increase the column flow to 3 or so and bakout the GC inlet and column as well.